move failed no model mag calibration in axis x
Added by James Bouwer almost 8 years ago
Hi Anchi,
I'm trying to calibrate goniomodeling on our Polara with energy filter in Leginon. You've used this system before. After the calibration, I use the navigator node and try to move using the model stage position and I get the error "move failed no model mag calibration in axis x". I've followed the instructions exactly a number of times. Is there something that I'm missing. Thanks.
Best Regards,
James
Replies (1)
RE: move failed no model mag calibration in axis x - Added by Anchi Cheng almost 8 years ago
Hi, James,
It is possible that the calibration was not completed with a complete model with both axes. Check in Presets Manager. Choose the preset, it should display the date the most recent calibration is done. Model calibration do need two axises measured and calibrated under the same label which in the first time is assigned to the session name but changeable in Advanced settings of the calibration node.
The other possibility is that your calibration ends up not in EF-TEM. It is hard for me to be sure on this, especially since you have tried multiple times. I can write a script to help you check if it is still not resolved.