pixel size calibration - other mags needed
Added by @Maria Janssen over 10 years ago
Hi, I was wondering how to adjust magnifications in the pixel size calibration table. We just installed EFTEM and all mags have changed. Is there a file I need to update?
Thanks,
Mandy
Replies (2)
RE: pixel size calibration - other mags needed - Added by Anchi Cheng over 10 years ago
I actually don't know how the scope behaves when EFTEM is added post Leginon installation. If they are all unique magnifications, Maybe you can update them in the Scope node on the scope PC, and then they will show up in Pixel Size calibration node in Calibration Application.
Before you try it, you should probably backup your database in case I am wrong.
RE: pixel size calibration - other mags needed - Added by @Maria Janssen over 10 years ago
Thanks so much! Yes, I got the new magnifications via the EM node and now they show up correctly in the pixel size calibration node.